Originally Posted by lbjvg
Perhaps there is too much back flow pressure from the other machine. The machines supposedly have built in back flow checks but perhaps it is not functional. You can purchase check valves that insert between the machine and the Y connector ( http://www.usplastic.com/catalog/cat...name=45&Page=1)
|
Yes it does appear the pressure/flow from one oxycon (probably the AirSep @ 6-7 psig) is back pressurizing the other oxycon (the Invacare Mobilaire V @ 5 psig).
The back flow check valve in the Invacare is probably working properly and is closing as it is designed to do due to back pressure from the AirSep. When closed, that same back flow check valve
also restricts oxygen outflow - thus reducing oxygen outflow to zero, or at the least, significantly restricting outflow from the oxycon to below the minimum alarm/shutdown tolerances (sometimes preset at <0.25 lpm flow), resulting in an alarm/shutdown.
Essentially, when the oxycon's back flow check valve closes, nothing goes into, and nothing comes out of, the oxycon's oxygen supply connection until the back pressure is removed - then the valve will automatically open. However, if the back pressure situation lasted "too long", the oxycon may have shutdown, and even with a automatic opening back flow check valve , no oxygen would flow from the shutdown oxycon.
CAVEAT
This applies to AirSep Elites and Invacare Mobilaire V oxycons. It
may or may not apply to other brands/models of oxycons, but I do not know that for certain.
Bill
p.s. I make no terminology distinction between concentrators, generators, etc.. To me they are all oxycons. They all concentrate oxygen. Some do a little - some do a lot - some do a whole lot. Some are old medical units - some are new - some are newly designed specifically for flamework (sort'a). But they all concentrate oxygen out of thin air.